    Since the patient has complete AV block and dependent most likely it is a PVC fused with paced complex, other possibility anodal capture although I do not think device is programmed to RV tip. What are your thoughts?

    • 1
    • Agree. In CRT whenever morphology changes you will consider loss of capture in one lead, especially left, but this happens in one beat and is still relatively narrow, so likely a fortuitous PVC fuses with paced beats.